Hi Charles,

Connecting an external power is possible but I am still lost to your requirement. Can you please explain in detail? What is the sensor you want to read and why does it need a long wait time? What is the sensor output that the logger needs to read?

Best regards,
Rudy Gunawan

Hi Charles, Connecting an external power is possible but I am still lost to your requirement. Can you please explain in detail? What is the sensor you want to read and why does it need a long wait time? What is the sensor output that the logger needs to read? Best regards, Rudy Gunawan

It's a HS 420 series accelerometer, which is actually a 4-20 mA velometer.

It's a HS 420 series accelerometer, which is actually a 4-20 mA velometer.

Hi Charles,

Since the sensor has 4-20 mA output and uses a junction box, I assume you try to utilize the internal shunt resistor. It would be neat on wiring, but because the internal shunt resistor is separated by relay, the logger would only close the relay when the measurement is due.

Obviously, because the shunt resistor is not connected to the 4-20 mA circuit, you may experience a required delay on measurement (long wait time). I recommend you to use an external 100 ohms shunt resistor instead and perhaps resistor network type to get the wiring neat for each terminal. Furthermore, if you utilize the external shunt resistor, you can connect more than one sensor per terminal.

Best regards,
dataTaker Expert

Hi Charles, Since the sensor has 4-20 mA output and uses a junction box, I assume you try to utilize the internal shunt resistor. It would be neat on wiring, but because the internal shunt resistor is separated by relay, the logger would only close the relay when the measurement is due. Obviously, because the shunt resistor is not connected to the 4-20 mA circuit, you may experience a required delay on measurement (long wait time). I recommend you to use an external 100 ohms shunt resistor instead and perhaps resistor network type to get the wiring neat for each terminal. Furthermore, if you utilize the external shunt resistor, you can connect more than one sensor per terminal. Best regards, dataTaker Expert
43
3
2
live preview
enter atleast 10 characters
WARNING: You mentioned %MENTIONS%, but they cannot see this message and will not be notified
Saving...
Saved
With selected deselect posts show selected posts
All posts under this topic will be deleted ?
Pending draft ... Click to resume editing
Discard draft