nseii posted Apr 13 '19 at 3:40 pm
I formatted a 4GB Kingston DataTraveler 100 thumb drive (USB 2.0) in a DT82E.

The Datataker read the USB for many minutes before returning front panel control.
prior to formatting it.
It took quite a while to format (no big surprise).
I then entered runjoboninsert"10OCTO15".
This worked and created the appropriate SNxxxxx directory with a job file in it.
I entered removemedia a
recent by admin  ·  Apr 13 '19 at 3:50 pm
32
2
nseii posted Apr 13 '19 at 3:37 pm
I have a DT82E that will not respond to a hard reset by pushing the reset button or a safe mode reset using the same reset button. The display is not showing any text. The sample and Data lights are blinking at the apparent same time 3 or for times a second and the power light is blinking at about the same rate but out of sync with the others.

I have attempt both in case there is something wr
recent by admin  ·  Apr 13 '19 at 3:47 pm
56
10
Hellou!

I having big problems when using Datataker to Loadcell measurements. Loadcell has been calibrated so, that when using 10 V excitation voltage the linear gauge factor is 99,453 kN/mV. Now we get numbers between 600-1000 mV and there are no scaling factor during measurement. Loadcell should give numbers between 0-2 mV, because the calibration area is that. What can be wrong? The loadcel
recent by admin  ·  Apr 12 '19 at 4:24 pm
34
4
yzak posted Apr 12 '19 at 4:01 pm
I was trying to setup an eagle.io account using the information from the link below
http://docs.eagle.io/en/latest/topics/device_configuration/datataker/index.html

I have finished the setup successfully but i couldn't get any data in my eagle account
In the datataker I have a schedule with 3 thermocouples connected that scans every 5 sec and an another schedule with 5 sec exclusively fo
recent by yzak  ·  Apr 12 '19 at 4:17 pm
6
1
sjankovi posted Apr 12 '19 at 3:59 pm
With DT80 G I have to measure time (let say in seconds) for which one digital input is in the status ON (how long it stays ON). Once when it goes OFF time has to be reset i.e. set to 0. It is not clear to me how to do this.

recent by admin  ·  Apr 12 '19 at 4:06 pm
24
1
basileJ posted Apr 12 '19 at 3:58 pm
Hello all,
I have some(many) "error" with reading of Vibrating Wires (range 600-1200Hz). i put a Measurement Delay but nothing seems to work well.
With manual reader (GK404 or sisgeo type) no problem appears, the value is stable. I use a DT80G serie3 and the measurement is very degraded , (sometimes i can read the real value same with my manual reader, some other times i read about 2600Hz
recent by admin  ·  Apr 12 '19 at 4:05 pm
12
1
Thaar posted Apr 12 '19 at 3:57 pm
welcome
we have DT85G recently, and we need to connect to it 3-wires LVDT of monitran company. we did several attempts of connection and all these gave an error reading.
the lvtd features are stroke 125mm , supply voltage =9 to 24 Volts Dc(regulated) and its connection details are RED for supply volts DC , Blue for 0 volts DC and green for output(0-5).
with my regards

recent by admin  ·  Apr 12 '19 at 4:04 pm
14
1
gpradipta posted Apr 12 '19 at 3:55 pm
Dear Sir,

we are having two nos. DT85GM to be configured with various sensors e.g. VW displacement sensor. VW pressure sensor and SDI12 Inclinometer.
only the SDI12 sensors are not communicating at all. always we are getting " not yet set" or no response from sensor in command window. we tried programming through
adding serial SDI12 sensor under schedule but same result. we tested the
recent by admin  ·  Apr 12 '19 at 4:03 pm
45
1
smilauer posted Apr 12 '19 at 3:54 pm
Hi,
I solved a problem on communication between DT80g (and likely also other DTx) and Linux OS using standard serial terminal. You can send text commands and receive output in the same way as with DeTransfer or in dEX command window but conveniently from Linux environment. This solution was tested on Ubuntu 16.04 but should work on other Linux-based systems as well. Here are the details.

S
5cb0a655e013d
recent by admin  ·  Apr 12 '19 at 4:03 pm
20
4
paddy posted Apr 12 '19 at 3:16 pm
I am looking for advice on an issue i am having with a DT82I it was setup to record data at 0.5 second intervals which was producing to many rows of information for excel to produce a spreadsheet of a weekly download which is how i was hoping it would work, subsequently i have a new oninsert file written and i thought that after inserting this file if i was to remove the usb clear it of files and
recent by paddy  ·  Apr 12 '19 at 3:38 pm
74
5
jianhui posted Apr 12 '19 at 3:18 pm
Hi,

Has anyone have experience on retrieving readings from DataTaker using Matlab?

I have open the ethernet port using the following:
t=tcpip('....IP ADDRESS...',7700)
fopen(t)

I have no idea how to continue when comes to retrieving the readings in MODBUS.

In the DT program, I have defined the following
1MODBUS(AD2,R3:29,=1..10CV) %Address 2, starting address 3000, regis
recent by jianhui  ·  Apr 12 '19 at 3:36 pm
25
3
ryan posted Apr 12 '19 at 3:19 pm
hi,

i am trying to upload my logged data direct to a USB. upon inserting the USB drive the datataker displays "reading USB" but does not begin writing.

i have read in the manual that you must have a file on the USB called ONINSERT.DXC

i have not found any information on how to create this file or where to find it on the supplied disk.

any assistance would be appreciated as i ha
recent by ryan  ·  Apr 12 '19 at 3:33 pm
18
1
Hi,

We have recently purchased a DT82I for use with an industrial coater. The coater sends strings of real time parameters which are forwarded through a RS232 port to a dot matrix printer. I was hoping to use the datalogger to retain a digital copy on a usb and then send the print command to the dot matrix printer. Since both printer and coater communicate only over RS232 I was planning on co
recent by Roger  ·  Apr 12 '19 at 3:21 pm
26
5
Hi,

I'm running a program to read and record values from an PT-100 RTD and a MODBUS device. The problem is, the resistance value of the PT-100 isn't read properly by the dataTaker. If I run the PT-100 program alone, there is no issue about the reading. But once the MODBUS device is connected to the sensor port, and the program runs together, the resistance values of the PT-100 increases by m
recent by admin  ·  Apr 12 '19 at 3:14 pm
18
3
arivan.g posted Apr 9 '19 at 1:04 pm
Hi,

I'm using the DT82I for logging current and voltage values from a current monitoring device. I've used around 170 CV's to log and display the data, in addition to individual labels. The data has also been collected without any problem.

The problem is that the last 10-15Cv's seem to have no label/name. I'm running similar programs on 2 separate DT82I's, and both have the same problem.
5cac8a346ef33
recent by Roger  ·  Apr 12 '19 at 3:10 pm
32
2
arivan.g posted Apr 9 '19 at 1:00 pm
Hi,

We've got a DT82I running 24/7 at a remote site. The program is meant to retain some CV's to be used as reference for the next day.

But, I've noticed that my DT82I sometimes doesn't seem to retain those values. Upon checking, we found that the DT82I displays the "DT80 restarted / Power loss" message every morning. But we have it running on an UPS, so I doubt power supply is an issue.
recent by arivan.g  ·  Apr 12 '19 at 3:07 pm
30
7
StephenR posted Apr 8 '19 at 3:15 pm
Hi,
I have a dt800 which i wish to connect a number of energy meters to. the energy meters have a pulse output of 100p/kWh. my intention is to use a counter channel on the dt800 to accept the pulse output from the meters and also send this reading to a channel variable. within another channel variable i intended to scale the input by the simple equation 1CV=2CV/100 which should give me kWh. i
recent by Roger  ·  Apr 9 '19 at 1:27 pm
21
1
Hi,

I'm trying to do manual data downloads via a 2 GB USB stick. However, every time I insert the USB stick, the logger runs into SW Exception error.

I have the latest 9.20.8942 firmware with a DT82I Series 3. The error I'm getting is Reset 9.20.8942 EXCEPTION (c8610007).

Is there a way to fix this issue? I need to resolve this ASAP as I will be deploying these into the field shortly.
recent by admin  ·  Apr 9 '19 at 1:14 pm
25
2
anupkc posted Apr 9 '19 at 12:42 pm
Hi, I am using datataker DT82I series 4. Previously, I was using deLogger but wanted to try out dEX.
I have connected the datataker to my computer via Ethernet cable and load dEX via on Mozilla Firefox (v61). When I create any schedule or add any calculations and try to type in the values, there is a lag. The characters appear on the screen 5-8 seconds later. It's so frustrating when I need to
recent by anupkc  ·  Apr 9 '19 at 12:57 pm
52
3
Chandran posted Apr 9 '19 at 12:46 pm
Hi,
Does anyone have experience using JBUS with Datataker? I appreciate if you can share your experience. I read that JBUS is similar to MODBUS except for the adressing. I found it is NOT true. Any comments?

recent by KadeM  ·  Apr 9 '19 at 12:56 pm
62
1
CTLHelpdesk posted Apr 9 '19 at 12:45 pm
Recently we received 2 of the DataTaker loggers.
The one I am writing about is the DT82I.
My manager recently received a firmware update that allowed the ability to setup security for separate log in's.
He was try to configure the device and ran out of time so I took over.
During my interaction with this device I have had it lock up several times and quite honestly had to pull the bat
recent by Roger  ·  Apr 9 '19 at 12:51 pm
41
1
Chandran posted Apr 9 '19 at 12:47 pm
Hi,
When there is no DNS server, the DT generate it's own IP adress 169.254.xxx.xxx. Is there any way we can use the key on the DT to change this address? I feel this way is easier than changing the computer's address to match DT's subnet. Any feed back is appreciated.

TQ [Chandran]

recent by Roger  ·  Apr 9 '19 at 12:51 pm
48
11
Skeighter posted Apr 8 '19 at 2:38 pm
Hello!

I currently transferring files eveary 30 min, with ftp protocol to my pc via DDNS ( with no-ip ). But i have notice that after a while DT800 blocks the transfers. The way i have to resolve is to make singlepush to DT800 and then the problem is gone. But i dont like to do that all the time.
Debuging DT800 while the problem is present, i check that i can still connect via TCP/IP, port
recent by admin  ·  Apr 8 '19 at 3:19 pm
31
4
Hello,

I have set up a current-excited bridge (BGI) circuit for a DT800 to measure the strain across five quarter-bridges on a channel pair, as shown in the attachment. The problem I am having is that I do not know what scaling factor to use to convert the incoming ppm signal to microstrain for each measurement. In addition, I am not always running all five bridges, which I assume changes the
5cab540d144c6
recent by Roger  ·  Apr 8 '19 at 3:02 pm
16
2
alexbe posted Apr 8 '19 at 2:33 pm
My DT800 has suddenly started giving me incorrect outputs around 10% low. Inputs are 6 LVDT and 1 PT100 temperature probe, all the used channels are incorrect. LVDT typically over a scaled 30mm range reads around 28.4mm and the PT100 reads around 95°C for an actual temperature of 106°C, and the outputs are not stable. Software is a customised version of DE logger written specifically for
recent by alexbe  ·  Apr 8 '19 at 2:53 pm
20
2
fredflinston posted Apr 8 '19 at 2:32 pm
I have to use GPS18x 5Hz for some position/speed measurements. Want to use DT800 serial channel for that.
I am sure some years ago (ie in the old forum) there was some discussion on this topic, with programming sample(s).

I am writing a short program using NMEA sentences, but must be doing something wrong for have no joy!!!!!!!
Would anyone have any relevant information, please???

10
1
baybars posted Apr 8 '19 at 2:35 pm
Hello,

I am trying to use ICP type of Accelerometer with Datataker DT800.

For this I need 4mA constant current supply continuously from Datataker.

However, datataker does not supply (or I couldn't adjust it) current continuously. It excites when it measures.

Is there any way to supply current without break?

Thanks

recent by admin  ·  Apr 8 '19 at 2:50 pm