I'm experimentally studying the self-heating of a special type of resistors. Typically I'm using a DT800 to record differential DC voltages as a "V" channel between, for example, 1+ and 1- terminals, or 3+ and 3-, et cetera, in a simple resistor network. In my experiment I gradually reduce the power supply voltage driving the entire network, over time. Unfortunately my approach is fairly vulnerable to the measured voltages approaching something a little different from zero, as the true physical voltages approach zero. A certain figure of merit in my system varies as the cube of the error in measuring true zero voltages. I get less than a millivolt error when the true voltage is zero, but if practical I'd like better. The data sheet says there are 12 auto-scaling ranges starting at 10 mV full scale, so I think the least significant of the 16 bits should represent a third of a microvolt and I'm seeing a hundred times that or so as my error. Is that to be expected?

Therefore I started some side experiments with zero voltages on the DT800. I am posting because I'd like more insight into what I'm seeing.

When I daisy-chain a bus bar wire across 1+, 1-, 2+, 2-, 3+, 3-, 4+, and 4-, and also tie it to Ch or Sr, I believe each of these 4 differential inputs is physically getting a true zero voltage, so ideally 1V, 2V, 3V and 4V should read 0 mV. I do averaging within records like this:
RS01T ' Subsampling at 100 Hz for later averaging
RA1S ' Schedule "A" runs at 1 Hz
and I run this overnight or over the weekend to get lots of points.

Oddly, the 1V channel reads about 0.05 to 0.1 mV, the 2V channel reads about 0.01 mV, the 3V channel reads about 0.003 mV, and the 4V channel reads about 0.0008 mV. That is, each channel number is several times more accurate than the channel number below it. It isn't just a fluke. I'm doing this with four different DT800 loggers (SN082868, SN082893, SN092777, and SN094769) and they all do this. Can anybody give me insight as to what is going on here?

In these results, the range of values I get excludes 0. That is, a measurement of 0 is many many standard deviations away from the mean.

I notice that sometimes I see a pattern that looks very much like the HVAC cycling in my lab space, including the cutback feature that sets less stringent temperatures for overnight or weekend periods. From the manual notes around page 188 about the autozero correction I understand that changing temperature is a possible source of error.

Also, I wondered if input impedance limited accuracy, so I set one DT800 up with its inputs shorted with a 100k resistor, a 1M, a 10M, a 100K, 1M, 10M and also a wire short. I tied the - side of all these inputs together to Ch. There is no tendency for the higher impedances to correspond to higher (or different) errors.

I experimented a little with the VNC channel type and learned that the autozero correction makes things way better. All my above results are with the V channel type which includes autozero correction.

Perhaps the reading accuracy is best at the ends of the scale, which in the DT800 are at plus and minus voltages, never at 0. I might be able to bias my physical experiment to put the readings at the low end of the measurement scale, so I'm measuring there and not in the middle of the scale. Is there an easy way to do that?

I've seen a bunch of other things but haven't spotted a pattern or made sense of it. Any help you might give me in understanding this would be very appreciated.

Thank you!

P.S. I guess I should ask, is there a market for antique DT800 loggers?

I'm experimentally studying the self-heating of a special type of resistors. Typically I'm using a DT800 to record differential DC voltages as a "V" channel between, for example, 1+ and 1- terminals, or 3+ and 3-, et cetera, in a simple resistor network. In my experiment I gradually reduce the power supply voltage driving the entire network, over time. Unfortunately my approach is fairly vulnerable to the measured voltages approaching something a little different from zero, as the true physical voltages approach zero. A certain figure of merit in my system varies as the cube of the error in measuring true zero voltages. I get less than a millivolt error when the true voltage is zero, but if practical I'd like better. The data sheet says there are 12 auto-scaling ranges starting at 10 mV full scale, so I think the least significant of the 16 bits should represent a third of a microvolt and I'm seeing a hundred times that or so as my error. Is that to be expected? Therefore I started some side experiments with zero voltages on the DT800. I am posting because I'd like more insight into what I'm seeing. When I daisy-chain a bus bar wire across 1+, 1-, 2+, 2-, 3+, 3-, 4+, and 4-, and also tie it to Ch or Sr, I believe each of these 4 differential inputs is physically getting a true zero voltage, so ideally 1V, 2V, 3V and 4V should read 0 mV. I do averaging within records like this: RS01T ' Subsampling at 100 Hz for later averaging RA1S ' Schedule "A" runs at 1 Hz and I run this overnight or over the weekend to get lots of points. Oddly, the 1V channel reads about 0.05 to 0.1 mV, the 2V channel reads about 0.01 mV, the 3V channel reads about 0.003 mV, and the 4V channel reads about 0.0008 mV. That is, each channel number is several times more accurate than the channel number below it. It isn't just a fluke. I'm doing this with four different DT800 loggers (SN082868, SN082893, SN092777, and SN094769) and they all do this. Can anybody give me insight as to what is going on here? In these results, the range of values I get excludes 0. That is, a measurement of 0 is many many standard deviations away from the mean. I notice that sometimes I see a pattern that looks very much like the HVAC cycling in my lab space, including the cutback feature that sets less stringent temperatures for overnight or weekend periods. From the manual notes around page 188 about the autozero correction I understand that changing temperature is a possible source of error. Also, I wondered if input impedance limited accuracy, so I set one DT800 up with its inputs shorted with a 100k resistor, a 1M, a 10M, a 100K, 1M, 10M and also a wire short. I tied the - side of all these inputs together to Ch. There is no tendency for the higher impedances to correspond to higher (or different) errors. I experimented a little with the VNC channel type and learned that the autozero correction makes things way better. All my above results are with the V channel type which includes autozero correction. Perhaps the reading accuracy is best at the ends of the scale, which in the DT800 are at plus and minus voltages, never at 0. I might be able to bias my physical experiment to put the readings at the low end of the measurement scale, so I'm measuring there and not in the middle of the scale. Is there an easy way to do that? I've seen a bunch of other things but haven't spotted a pattern or made sense of it. Any help you might give me in understanding this would be very appreciated. Thank you! P.S. I guess I should ask, is there a market for antique DT800 loggers?

More details and some corrections:

Here is the study of loggers with their inputs wired short. I alluded to these values earlier. These means and standard deviations are all in millivolts, so a few microvolts to a few dozen microvolts.
Four Loggers with Shorted Inputs
Sn Ch Mean StdDev
SN082868 5 0.082902 .002585310
SN082868 6 0.009542 .001014568
SN082868 7 0.003685 .000976722
SN082868 8 0.001171 .000937215
SN082893 5 0.079420 .002002098
SN082893 6 0.009579 .000899322
SN082893 7 0.003488 .000862541
SN082893 8 0.001307 .000844087
SN092777 5 0.055841 .001201452
SN092777 6 0.003152 .000999203
SN092777 7 0.000963 .000960528
SN092777 8 0.000721 .000938262
SN094769 5 0.098538 .002035279
SN094769 6 0.009004 .001306048
SN094769 7 0.003023 .001323405
SN094769 8 0.000972 .001351155

Here is the study of different load resistors shorting the inputs. This table shows means and standard deviations in millivolts, also, like above. So I'm getting values like -58 millivolts and +64 millivolts. Not microvolts. These are big zero measurements! I misspoke earlier. These values are many times higher than the ones above.

SN082893 Inputs shorted with 10^LogLoadRes ohm resistor
Ch Mean StdDev LogLoadRes
5 14.7109 0.14413 5
6 13.8488 0.14125 6
7 -57.9377 0.12537 7
8 2.4911 0.07505 5
9 64.0819 0.02583 6
10 64.0562 0.02590 7
11 0.3735 0.00057 -1

More details and some corrections: Here is the study of loggers with their inputs wired short. I alluded to these values earlier. These means and standard deviations are all in millivolts, so a few microvolts to a few dozen microvolts. Four Loggers with Shorted Inputs Sn Ch Mean StdDev SN082868 5 0.082902 .002585310 SN082868 6 0.009542 .001014568 SN082868 7 0.003685 .000976722 SN082868 8 0.001171 .000937215 SN082893 5 0.079420 .002002098 SN082893 6 0.009579 .000899322 SN082893 7 0.003488 .000862541 SN082893 8 0.001307 .000844087 SN092777 5 0.055841 .001201452 SN092777 6 0.003152 .000999203 SN092777 7 0.000963 .000960528 SN092777 8 0.000721 .000938262 SN094769 5 0.098538 .002035279 SN094769 6 0.009004 .001306048 SN094769 7 0.003023 .001323405 SN094769 8 0.000972 .001351155 Here is the study of different load resistors shorting the inputs. This table shows means and standard deviations in millivolts, also, like above. So I'm getting values like -58 millivolts and +64 millivolts. Not microvolts. These are big zero measurements! I misspoke earlier. These values are many times higher than the ones above. SN082893 Inputs shorted with 10^LogLoadRes ohm resistor Ch Mean StdDev LogLoadRes 5 14.7109 0.14413 5 6 13.8488 0.14125 6 7 -57.9377 0.12537 7 8 2.4911 0.07505 5 9 64.0819 0.02583 6 10 64.0562 0.02590 7 11 0.3735 0.00057 -1