Hi. I'm having trouble communicating with my DT800 s1 via Ethernet.

I have managed to connect via serial and I've ran through the setup procedure but it does not appear as a network location.

I'm running DeTransfer on a spare PC we have which is running Win 7 Home premium, I'm wondering if its something to do with this ver of windows as its networking capabilities are limited or maybe its something else?

any ideas anyone?

Thanks.

Hi. I'm having trouble communicating with my DT800 s1 via Ethernet. I have managed to connect via serial and I've ran through the setup procedure but it does not appear as a network location. I'm running DeTransfer on a spare PC we have which is running Win 7 Home premium, I'm wondering if its something to do with this ver of windows as its networking capabilities are limited or maybe its something else? any ideas anyone? Thanks.

Hi OEC,

You must set all Ethernet setting manually as DT800 can't get any DHCP assignment, example:
PROFILE"ETHERNET","UDP_SUPPORTED"="YES"
PROFILE"ETHERNET","IP_ADDRESS"="192.168.1.10"
PROFILE"ETHERNET","GATEWAY"="192.168.1.1"

Then use SINGLEPUSH command to reset the logger.
From DeTransfer you can set a connection to 192.168.1.10 on port 8 (remember the port number is different than DT80 series).

Best regards,
Rudy Gunawan

Hi OEC, You must set all Ethernet setting manually as DT800 can't get any DHCP assignment, example: PROFILE"ETHERNET","UDP_SUPPORTED"="YES" PROFILE"ETHERNET","IP_ADDRESS"="192.168.1.10" PROFILE"ETHERNET","GATEWAY"="192.168.1.1" Then use SINGLEPUSH command to reset the logger. From DeTransfer you can set a connection to 192.168.1.10 on port 8 (remember the port number is different than DT80 series). Best regards, Rudy Gunawan

Hi, thanks! I have it connected now.
I wonder if you could help, I'm trying to communicate with a load cell that i have connected.
I have connected the +VE & -VE supply to the * & # of ch1 and ive connected the signals wires to the + & -
Could you give me an example or direct me toward an example of the type of schedule script i could use to log data from the cell?

many thanks again smile

Hi, thanks! I have it connected now. I wonder if you could help, I'm trying to communicate with a load cell that i have connected. I have connected the +VE & -VE supply to the * & # of ch1 and ive connected the signals wires to the + & - Could you give me an example or direct me toward an example of the type of schedule script i could use to log data from the cell? many thanks again :smile:

Hi

Bridge connection is commonly used to read load cell

please see the below link for DT800 user Manual

http://www.thermofisher.com.au/Uploads/file/Environmental-Industrial/Process-Monitoring-Industrial-Instruments/Data-Acquisition/Data-Loggers/DataTaker/UM-0068-A2-DT800-User's-Manual.pdf

and refer to following section for wiring and configuration

Bridges
WIRING DIAGRAMS: see “Bridge Inputs” beginning on page 174

Thanks
Regards
Lokesh

Hi Bridge connection is commonly used to read load cell please see the below link for DT800 user Manual http://www.thermofisher.com.au/Uploads/file/Environmental-Industrial/Process-Monitoring-Industrial-Instruments/Data-Acquisition/Data-Loggers/DataTaker/UM-0068-A2-DT800-User's-Manual.pdf and refer to following section for wiring and configuration Bridges WIRING DIAGRAMS: see “Bridge Inputs” beginning on page 174 Thanks Regards Lokesh
35
4
3
live preview
enter atleast 10 characters
WARNING: You mentioned %MENTIONS%, but they cannot see this message and will not be notified
Saving...
Saved
With selected deselect posts show selected posts
All posts under this topic will be deleted ?
Pending draft ... Click to resume editing
Discard draft