dear Sir,
one of customers is using dt800 for resistance measurement.
recently we are facing the following problem;
measured resistance values are lower than actual e.g. 260 Ohm is measured while actual is 350 Ohm.
test command generated following:
T,081617,2014/04/10,12:34:42,0.205322,0;"dataTaker 800 Version 4.10.0001 Flash 2006/01/18 11:32:09";
T,081617,2014/04/10,12:34:42,0.205322,1;"081617";
T,081617,2014/04/10,12:34:42,0.205322,2;"None";
T,081617,2014/04/10,12:34:42,0.205322,3;"ATA flash card (61.11 MB)";
T,081617,2014/04/10,12:34:42,0.205322,4;17.1;
T,081617,2014/04/10,12:34:42,0.205322,5;13.5,1;
T,081617,2014/04/10,12:34:42,0.205322,6;16.5,1;
T,081617,2014/04/10,12:34:42,0.205322,7;0.6,0;
T,081617,2014/04/10,12:34:42,0.205322,8;8.9,0;
T,081617,2014/04/10,12:34:42,0.205322,9;-7.6,0;
T,081617,2014/04/10,12:34:42,0.205322,10;5.3,1;
T,081617,2014/04/10,12:34:42,0.205322,11;2.5,1;
T,081617,2014/04/10,12:34:42,0.205322,12;0.1,1;
T,081617,2014/04/10,12:34:42,0.205322,13;0.0,0;
T,081617,2014/04/10,12:34:42,0.205322,14;0;
Lithium Volts; 0.6 V FAIL
Analig pos supply: 8.9V FAIL
Analog neg Supply: -7.6V FAIL
CMRR : 0.0 FAIL
Total Health ; FAIL
pl advise,
regards,
Pradipta
dear Sir,
one of customers is using dt800 for resistance measurement.
recently we are facing the following problem;
measured resistance values are lower than actual e.g. 260 Ohm is measured while actual is 350 Ohm.
test command generated following:
T,081617,2014/04/10,12:34:42,0.205322,0;"dataTaker 800 Version 4.10.0001 Flash 2006/01/18 11:32:09";
T,081617,2014/04/10,12:34:42,0.205322,1;"081617";
T,081617,2014/04/10,12:34:42,0.205322,2;"None";
T,081617,2014/04/10,12:34:42,0.205322,3;"ATA flash card (61.11 MB)";
T,081617,2014/04/10,12:34:42,0.205322,4;17.1;
T,081617,2014/04/10,12:34:42,0.205322,5;13.5,1;
T,081617,2014/04/10,12:34:42,0.205322,6;16.5,1;
T,081617,2014/04/10,12:34:42,0.205322,7;0.6,0;
T,081617,2014/04/10,12:34:42,0.205322,8;8.9,0;
T,081617,2014/04/10,12:34:42,0.205322,9;-7.6,0;
T,081617,2014/04/10,12:34:42,0.205322,10;5.3,1;
T,081617,2014/04/10,12:34:42,0.205322,11;2.5,1;
T,081617,2014/04/10,12:34:42,0.205322,12;0.1,1;
T,081617,2014/04/10,12:34:42,0.205322,13;0.0,0;
T,081617,2014/04/10,12:34:42,0.205322,14;0;
Lithium Volts; 0.6 V FAIL
Analig pos supply: 8.9V FAIL
Analog neg Supply: -7.6V FAIL
CMRR : 0.0 FAIL
Total Health ; FAIL
pl advise,
regards,
Pradipta