Hi Roger and Rudy,
Thank you for the responses.
I did a simple experiment to test what the gage factor should be, using a beam with a mass on the end to cause bending which results in a known strain.
I then set up my circuit as shown before and measured the strain whilst varying the amount of 1/2 bridges (with one active gage each) on the circuit as well as changing between voltage excitation and current excitation. I have attached my results with my hand calculations. The gage factor is 2.08.
From the results it can be seen that the voltage excitation measurement is not affected by the added 1/2 bridges however the current excitation measurement is.
For the current excitation method the excitation voltage should decrease by a factor of [(the number of 1/2 bridges in the circuit, including the completion 1/2 bridge)/2] when adding additional 1/2 bridges to the circuit, shown in attached calculations.
Therefore, for one active 1/2 bridge the voltage reduction is 2/2, for two active 1/2 bridges it is 3/2, for three active 1/2 bridges it is 4/2 and so on. These factors are the same as the factors which the measured strain is out by for the current excitation method.
However, the output voltage is relative to the excitation voltage, so it should be reduced by the same factor. Therefore, the measured strain should not be affected as the reduced voltage factor would cancel out, as shown in calculations.
I have checked my circuit and done the test twice over, so I am not sure why the current excitation method is causing these results?
Thank you,
Matt
Hi Roger and Rudy,
Thank you for the responses.
I did a simple experiment to test what the gage factor should be, using a beam with a mass on the end to cause bending which results in a known strain.
I then set up my circuit as shown before and measured the strain whilst varying the amount of 1/2 bridges (with one active gage each) on the circuit as well as changing between voltage excitation and current excitation. I have attached my results with my hand calculations. The gage factor is 2.08.
From the results it can be seen that the voltage excitation measurement is not affected by the added 1/2 bridges however the current excitation measurement is.
For the current excitation method the excitation voltage should decrease by a factor of [(the number of 1/2 bridges in the circuit, including the completion 1/2 bridge)/2] when adding additional 1/2 bridges to the circuit, shown in attached calculations.
Therefore, for one active 1/2 bridge the voltage reduction is 2/2, for two active 1/2 bridges it is 3/2, for three active 1/2 bridges it is 4/2 and so on. These factors are the same as the factors which the measured strain is out by for the current excitation method.
However, the output voltage is relative to the excitation voltage, so it should be reduced by the same factor. Therefore, the measured strain should not be affected as the reduced voltage factor would cancel out, as shown in calculations.
I have checked my circuit and done the test twice over, so I am not sure why the current excitation method is causing these results?
![5cab53a29c53a](serve/attachment&path=5cab53a29c53a)
![5cab53a8bdf8a](serve/attachment&path=5cab53a8bdf8a)
![5cab53a3c604c](serve/attachment&path=5cab53a3c604c)
Thank you,
Matt