I would start by testing the strain gauge installation;
1/ Get a strain gauge test unit and check the resistance to ground. If you don't have very high resistance to ground then the gauge has been damaged and needs to be replaced. If the gauges the gauge needs to be in a moist environment (Embedded in concrete for example) then check the environmental coating in the same way.
2/ Check the quality of the soldering. This may require using a microscope. Check for dry joints, solder causing nearly causing short circuits or close to edges of the gauge. Your solder joints should be small round pillows. Also make sure you have removed all solder flux from the joints.
3/ Check your wiring. Make sure all ends of leads are fully enclosed in the solder. Where possible use shielded twisted pair cables and only ground the shielding at one end. The end that isn't grounded check the shield is covered by the insulation and not touching or close to the other surface. I go to the extent trimming the shield cable as close to the insulation as possible the of very lightly stretching the cable insulation just enough to cover any exposed shield. Then apply the environmental coating to the end of the cable.
4/ Check your cables don't run near any power supplies or power cables.
Cheers
Roger
Good morning Eloi,
I would start by testing the strain gauge installation;
1/ Get a strain gauge test unit and check the resistance to ground. If you don't have very high resistance to ground then the gauge has been damaged and needs to be replaced. If the gauges the gauge needs to be in a moist environment (Embedded in concrete for example) then check the environmental coating in the same way.
2/ Check the quality of the soldering. This may require using a microscope. Check for dry joints, solder causing nearly causing short circuits or close to edges of the gauge. Your solder joints should be small round pillows. Also make sure you have removed all solder flux from the joints.
3/ Check your wiring. Make sure all ends of leads are fully enclosed in the solder. Where possible use shielded twisted pair cables and only ground the shielding at one end. The end that isn't grounded check the shield is covered by the insulation and not touching or close to the other surface. I go to the extent trimming the shield cable as close to the insulation as possible the of very lightly stretching the cable insulation just enough to cover any exposed shield. Then apply the environmental coating to the end of the cable.
4/ Check your cables don't run near any power supplies or power cables.
Cheers
Roger