So the load cell is wired into the base and when you swap over the DT800 with another you have noise on one but not on the second DT800..
Yes.. but it's a strain gauge input card with 10 or so strain gauges. I use the externally powered five half bridges on a channel pair method.
When you say one has more noise than the other, how much more noise are we talking about?
About +/- 500 microstrain.... as in BIG! It gets worse when unplugging the power to the laptop making me think an earth loop problem in the logger - inducing a 3000uE offset. I might ad doing the same with the "good" 2nd logger has no effect on data.
Analog inputs in the 0-10 Volt range using V(#) seem to be ok on the "bad" logger while the BGV inputs are noisy.
Both passed the "test" command and were sent the same program - the problematic logger has worked ok until just recently until an away job making me think a "knock" by your average baggage handler has caused a fault.
Mitch
So the load cell is wired into the base and when you swap over the DT800 with another you have noise on one but not on the second DT800..
Yes.. but it's a strain gauge input card with 10 or so strain gauges. I use the externally powered five half bridges on a channel pair method.
When you say one has more noise than the other, how much more noise are we talking about?
About +/- 500 microstrain.... as in BIG! It gets worse when unplugging the power to the laptop making me think an earth loop problem in the logger - inducing a 3000uE offset. I might ad doing the same with the "good" 2nd logger has no effect on data.
Analog inputs in the 0-10 Volt range using V(#) seem to be ok on the "bad" logger while the BGV inputs are noisy.
Both passed the "test" command and were sent the same program - the problematic logger has worked ok until just recently until an away job making me think a "knock" by your average baggage handler has caused a fault.
Mitch