I don't use the analog output at all. Simply a battery and voltage divider. Cheap simple and quick. A thermocouple is just a voltage input and that's what it simulates.
Ok, that's what I thought initially but wasn't sure.
You could use a voltage divider to reduce the signal from the analog output. The analog output is 12 Bit and the range is +/- 10 VDC so the resolution it larger that what you would like for a thermocouple.
Yes, I've thought about a voltage divider today, and I think the resolution will be sufficient. If I bring the 10V down to 10mV then the resolution should be 0.010mV, which is roughly 2 or 3 degrees for a type-R thermocouple.
What are you trying to do???
I'm trying to use type-N thermocouples, convert the temperature reading to a type-R voltage within the DT800, and then use the Ao to feed a furnace controller that is expecting type-R thermocouple input.
Given that I would need approx 4 m of type-R to do this, at $800/m, as opposed to about $20/m for type-N, a few resistors on a breadboard are very attractive! The ability to massage the temp data within the DT800 before passing it to the furnace controller is an added bonus.
Thinking about this a bit, some builtin functions for converting temperatures to various type thermocouple voltages (assuming an external voltage divider) would be a great feature for the next firmware update!
I hope that explains it
Cheers,
Chris.
> I don't use the analog output at all. Simply a battery and voltage divider. Cheap simple and quick. A thermocouple is just a voltage input and that's what it simulates.
Ok, that's what I thought initially but wasn't sure.
> You could use a voltage divider to reduce the signal from the analog output. The analog output is 12 Bit and the range is +/- 10 VDC so the resolution it larger that what you would like for a thermocouple.
Yes, I've thought about a voltage divider today, and I think the resolution will be sufficient. If I bring the 10V down to 10mV then the resolution should be 0.010mV, which is roughly 2 or 3 degrees for a type-R thermocouple.
> What are you trying to do???
I'm trying to use type-N thermocouples, convert the temperature reading to a type-R voltage within the DT800, and then use the Ao to feed a furnace controller that is expecting type-R thermocouple input.
Given that I would need approx 4 m of type-R to do this, at $800/m, as opposed to about $20/m for type-N, a few resistors on a breadboard are very attractive! The ability to massage the temp data within the DT800 before passing it to the furnace controller is an added bonus.
Thinking about this a bit, some builtin functions for converting temperatures to various type thermocouple voltages (assuming an external voltage divider) would be a great feature for the next firmware update!
I hope that explains it :smile:
Cheers,
Chris.