Hey wassup.

I'm trying to measure frequency from a flow meters. So far I'm just trying to measure the freq from a function generator, I select a square wave of around 400 Hz and +-10V of amplitude and I connect it to channel 3.

I type "RA1s 3F" and the only result I get is -99999.9. I tried changing the threshold as indicated in the manual but I always the that same wrong result. How could I fix this??? Thanks a lot for your fast reply.

Hey wassup. I'm trying to measure frequency from a flow meters. So far I'm just trying to measure the freq from a function generator, I select a square wave of around 400 Hz and +-10V of amplitude and I connect it to channel 3. I type "RA1s 3F" and the only result I get is -99999.9. I tried changing the threshold as indicated in the manual but I always the that same wrong result. How could I fix this??? Thanks a lot for your fast reply.

Good morning toniojesusde,

Three points to check:

  1. Lock the gain to the 20 VDC range.
  2. The threshold level in in mV not Volts
  3. Do you have a ground connection to the AC terminal?

If you are measuring a flow meter you will be better off using the high speed counters rather than a frequency input.

Cheers,
Roger

Good morning toniojesusde, Three points to check: 1. Lock the gain to the 20 VDC range. 2. The threshold level in in mV not Volts 3. Do you have a ground connection to the AC terminal? If you are measuring a flow meter you will be better off using the high speed counters rather than a frequency input. Cheers, Roger
39
1
2
live preview
enter atleast 10 characters
WARNING: You mentioned %MENTIONS%, but they cannot see this message and will not be notified
Saving...
Saved
With selected deselect posts show selected posts
All posts under this topic will be deleted ?
Pending draft ... Click to resume editing
Discard draft