Wrong headers DT80 and CEM20

I have connected a CEM20 to a DT80W, using USB to download data the headers are completely wrong for the channels.
I have 101*TT("Temp 1") through to
105-TT("Temp14")

I have the CEM plugged into channel 1 on the DT.

The downloaded data in csv file gives me
Therm 1, Therm 1, Therm 1.........and so on. Including the VEXT channel

When unload through delogger5 headers are correct, what settings could cause this

I have tried reset and factorydefualts commands, no change.

The USB dcx file I added format=free and it downloaded a text file in the expected manner with correct channel names.

Also the CEM seems to poll all channels at once and not cycling through each channel like it used to. This is the first time using this logger and CEM had been used since it had returned from repairs.
Thanks in advance.
Damien.

I have connected a CEM20 to a DT80W, using USB to download data the headers are completely wrong for the channels. I have 101*TT("Temp 1") through to 105-TT("Temp14") I have the CEM plugged into channel 1 on the DT. The downloaded data in csv file gives me Therm 1, Therm 1, Therm 1.........and so on. Including the VEXT channel When unload through delogger5 headers are correct, what settings could cause this I have tried reset and factorydefualts commands, no change. The USB dcx file I added format=free and it downloaded a text file in the expected manner with correct channel names. Also the CEM seems to poll all channels at once and not cycling through each channel like it used to. This is the first time using this logger and CEM had been used since it had returned from repairs. Thanks in advance. Damien.
edited Jul 15 '21 at 10:50 am

Hi Wolfy27,

Please try to reinstall the firmware; there is something in the firmware that is causing the error (corrupted). If you have the 9.20 version, please upgrade it to 9.22.

If the upgrade does not work, please reformat the memory first using a standard command of the logger FORMAT B: DELETEALL.

Best regards,
dataTaker Expert

Hi Wolfy27, Please try to reinstall the firmware; there is something in the firmware that is causing the error (corrupted). If you have the 9.20 version, please upgrade it to 9.22. If the upgrade does not work, please reformat the memory first using a standard command of the logger FORMAT B: DELETEALL. Best regards, dataTaker Expert
72
2
2
live preview
enter atleast 10 characters
WARNING: You mentioned %MENTIONS%, but they cannot see this message and will not be notified
Saving...
Saved
With selected deselect posts show selected posts
All posts under this topic will be deleted ?
Pending draft ... Click to resume editing
Discard draft