Values over range and failed test report, Shunt(100R)

Hi

I have suddenly encountered problems with the readings from our DT85G. The temperatures readings we get is 327.7 degC, or OverRange.
When I run a test I get the following report below.
The Shunt(100R) gives a value of 0.0 Ohm.
I now wonder if and how this can be repaired and if a lightning strike could have cause this error?

Best regards
Josef

TEST report generated at 2021/07/27,09:58:51
dataTaker 85 Version 9.22.9125 Flash 2018/12/11 12:35:21

Product: DT85G-4
Serial Number: 110724

VEXT 11.8 V
VBAT (6V) 5.99 V PASS
IBAT +0 mA PASS
VSYS 7.09 V PASS
VLITH (3.6V) 3.63 V PASS
VDD (3.3V) 3.26 V PASS
VANA (3.8V) 3.62 V PASS
VRELAY (3.8V) 3.78 V PASS

VREF (2.5V) 2499.92 mV PASS
Ics I 0.86545 mA FAIL
Ics II 8.7028 mA FAIL
Ics III 0.86546 mA FAIL
Vos diff 0.0 uV PASS
Vos 3W I 0.1 uV PASS
Vos 3W II 11.0 uV PASS
Vos shunt 0.3 uV PASS
Vos + 0.6 uV PASS
Vos - 0.0 uV PASS
Vos -0.3 uV PASS
Vos # 0.0 uV PASS
Vos diff atten -0.3 uV PASS
Vos + atten 0.0 uV PASS
Vos - atten 0.0 uV PASS
Vos
atten 0.0 uV PASS
Term. factor 1.00470 PASS
Shunt (100R) 0.001 Ohm FAIL
CMRR 138.7 dB PASS

Overall health FAIL

Hi I have suddenly encountered problems with the readings from our DT85G. The temperatures readings we get is 327.7 degC, or OverRange. When I run a test I get the following report below. The Shunt(100R) gives a value of 0.0 Ohm. I now wonder if and how this can be repaired and if a lightning strike could have cause this error? Best regards Josef TEST report generated at 2021/07/27,09:58:51 dataTaker 85 Version 9.22.9125 Flash 2018/12/11 12:35:21 Product: DT85G-4 Serial Number: 110724 VEXT 11.8 V VBAT (6V) 5.99 V PASS IBAT +0 mA PASS VSYS 7.09 V PASS VLITH (3.6V) 3.63 V PASS VDD (3.3V) 3.26 V PASS VANA (3.8V) 3.62 V PASS VRELAY (3.8V) 3.78 V PASS VREF (2.5V) 2499.92 mV PASS Ics I 0.86545 mA FAIL Ics II 8.7028 mA FAIL Ics III 0.86546 mA FAIL Vos diff 0.0 uV PASS Vos 3W I 0.1 uV PASS Vos 3W II 11.0 uV PASS Vos shunt 0.3 uV PASS Vos + 0.6 uV PASS Vos - 0.0 uV PASS Vos * -0.3 uV PASS Vos # 0.0 uV PASS Vos diff atten -0.3 uV PASS Vos + atten 0.0 uV PASS Vos - atten 0.0 uV PASS Vos * atten 0.0 uV PASS Term. factor 1.00470 PASS Shunt (100R) 0.001 Ohm FAIL CMRR 138.7 dB PASS Overall health FAIL

Hi josefj,

The logger got a voltage surge from one or more connected sensors; you can send the unit back to us for repair. Would you please get in touch with your distributor for the return arrangement?

Please note, if you wish to use the logger, you will have a limited option. The faults are related to current excitation and internal shunt; anything without current excitation or internal shunt should still be OK.

Best regards,
dataTaker Expert

Hi josefj, The logger got a voltage surge from one or more connected sensors; you can send the unit back to us for repair. Would you please get in touch with your distributor for the return arrangement? Please note, if you wish to use the logger, you will have a limited option. The faults are related to current excitation and internal shunt; anything without current excitation or internal shunt should still be OK. Best regards, dataTaker Expert