General Discussions
Thermocouple Measurement Uncertainty

I am trying to calculate the measurement uncertainty of the system. I can't find any info in the manual. Does DT80 series 4 has any specified accuracy for thermocouple measurement or the internal cold junction compensation?

I am trying to calculate the measurement uncertainty of the system. I can't find any info in the manual. Does DT80 series 4 has any specified accuracy for thermocouple measurement or the internal cold junction compensation?

Hi Chance92,

dataTaker does not use the cold junction method since it is not applicable for portable loggers; however, we replace it by measuring the temperature on the board track using an RTD. The channel that is known as REFT is a temperature measurement coming from the RTD. If you can see the reading output from REFT, you should be able to read the thermocouple.

The accuracy for thermocouples in dataTaker is calculated through voltage measurement accuracy. You can find the formula in the DT8x user manual in the last section Q. That formula is only a guide on the worst-case situation, but you need to run the calculation based on the target temperature you want to read. At least for comparison.

The actual voltage reading accuracy can be seen from the calibration certificate, roughly 10 - 20% of the accuracy value from the calculation (5 to 10 times better than the calculation).

From the voltage reading accuracy, you can recalculate back to temperature using the thermocouple RvT table. And you will get the accuracy from the reading, please also consider the thermocouple accuracy as a sensor.

Best regards,
dataTaker Expert

Hi Chance92, dataTaker does not use the cold junction method since it is not applicable for portable loggers; however, we replace it by measuring the temperature on the board track using an RTD. The channel that is known as REFT is a temperature measurement coming from the RTD. If you can see the reading output from REFT, you should be able to read the thermocouple. The accuracy for thermocouples in dataTaker is calculated through voltage measurement accuracy. You can find the formula in the DT8x user manual in the last section Q. That formula is only a guide on the worst-case situation, but you need to run the calculation based on the target temperature you want to read. At least for comparison. The actual voltage reading accuracy can be seen from the calibration certificate, roughly 10 - 20% of the accuracy value from the calculation (5 to 10 times better than the calculation). From the voltage reading accuracy, you can recalculate back to temperature using the thermocouple RvT table. And you will get the accuracy from the reading, please also consider the thermocouple accuracy as a sensor. Best regards, dataTaker Expert
64
1
2
live preview
enter atleast 10 characters
WARNING: You mentioned %MENTIONS%, but they cannot see this message and will not be notified
Saving...
Saved
With selected deselect posts show selected posts
All posts under this topic will be deleted ?
Pending draft ... Click to resume editing
Discard draft