Support Forums
External power

Hi, I am trying to power the datataker externally using a 12V battery. I hooked it up on the +/- terminals on the side of the datataker, right next to the little BC bridge. It appeared to be working (I had the leads connected correctly) and charging the internal battery. However, one of the leads of the external 12V battery got disconnected and the internal battery was rapidly discharged and destroyed. I also noticed that just having the leads attached (i.e., without either being connected to the 12V battery) on the +/- terminal appears to drain the internal battery.

I would like to keep the internal battery connected as well so that I can occasionally disconnect the 12V battery and recharge it in my office without data loss. Is there a way/setting to prevent the 6V battery from discharging (other than ensuring leads do not get disconnected)? I do not want the internal 6V battery charging the 12V external battery.

Astrid

Hi, I am trying to power the datataker externally using a 12V battery. I hooked it up on the +/- terminals on the side of the datataker, right next to the little BC bridge. It appeared to be working (I had the leads connected correctly) and charging the internal battery. However, one of the leads of the external 12V battery got disconnected and the internal battery was rapidly discharged and destroyed. I also noticed that just having the leads attached (i.e., without either being connected to the 12V battery) on the +/- terminal appears to drain the internal battery. I would like to keep the internal battery connected as well so that I can occasionally disconnect the 12V battery and recharge it in my office without data loss. Is there a way/setting to prevent the 6V battery from discharging (other than ensuring leads do not get disconnected)? I do not want the internal 6V battery charging the 12V external battery. Astrid

This is the output when I connect a fully charged external battery, I replaced both lithium ion and internal battery 2 months ago:

TEST report generated at 2016/06/22,10:02:25
dataTaker 80 Version 9.18.6291 Flash 2015/08/07 14:28:13

Product: DT80-2
Serial Number: 086600

VEXT 12.7 V
VBAT (6V) 6.30 V PASS
IBAT +519 mA PASS
VSYS 7.12 V PASS
VLITH (3.6V) 3.68 V PASS
VDD (3.3V) 3.28 V PASS
VANA (3.8V) 3.57 V PASS
VRELAY (4.5V) 4.48 V PASS

VREF (2.5V) 2500.26 mV PASS
Ics I 0.21272 mA PASS
Ics II 2.5649 mA PASS
Vos diff 4.9 uV PASS
Vos 3W I 292.1 uV PASS
Vos 3W II -112.3 uV PASS
Vos shunt -30.5 uV PASS
Vos + 3.5 uV PASS
Vos - 2.8 uV PASS
Vos 4.5 uV PASS
Vos # -60.7 uV PASS
Vos diff atten 61.3 uV PASS
Vos + atten 61.1 uV PASS
Vos - atten 60.9 uV PASS
Vos
atten 61.8 uV PASS
Term. factor 1.00470 PASS
Shunt (100R) 99.826 Ohm PASS
CMRR 100.7 dB PASS

Overall health PASS

This is the output when I connect a fully charged external battery, I replaced both lithium ion and internal battery 2 months ago: TEST report generated at 2016/06/22,10:02:25 dataTaker 80 Version 9.18.6291 Flash 2015/08/07 14:28:13 Product: DT80-2 Serial Number: 086600 VEXT 12.7 V VBAT (6V) 6.30 V PASS IBAT +519 mA PASS VSYS 7.12 V PASS VLITH (3.6V) 3.68 V PASS VDD (3.3V) 3.28 V PASS VANA (3.8V) 3.57 V PASS VRELAY (4.5V) 4.48 V PASS VREF (2.5V) 2500.26 mV PASS Ics I 0.21272 mA PASS Ics II 2.5649 mA PASS Vos diff 4.9 uV PASS Vos 3W I 292.1 uV PASS Vos 3W II -112.3 uV PASS Vos shunt -30.5 uV PASS Vos + 3.5 uV PASS Vos - 2.8 uV PASS Vos * 4.5 uV PASS Vos # -60.7 uV PASS Vos diff atten 61.3 uV PASS Vos + atten 61.1 uV PASS Vos - atten 60.9 uV PASS Vos * atten 61.8 uV PASS Term. factor 1.00470 PASS Shunt (100R) 99.826 Ohm PASS CMRR 100.7 dB PASS Overall health PASS

Hi Astrid,

Internal battery has 1.2Ah in size, therefore depending on your measurement interval and number of channels it may get quickly drained.
Power consumption list in DT manual specific to battery usage has an assumption of enabling sleep.
Example:
5 sensors at 10 minutes interval consume 4 mW (according to the manual) because sleep mode is enable - internal battery will last for 1700+ hours
When sleep mode is disable it will consume 302 mW - internal battery will last for 23 hours.
The above number with an assumption Ethernet, LCD backlight are disable and no usage on 12V onboard output.

Could you describe your sensors and measurement interval?

Best regards,
Rudy Gunawan

Hi Astrid, Internal battery has 1.2Ah in size, therefore depending on your measurement interval and number of channels it may get quickly drained. Power consumption list in DT manual specific to battery usage has an assumption of enabling sleep. Example: 5 sensors at 10 minutes interval consume 4 mW (according to the manual) because sleep mode is enable - internal battery will last for 1700+ hours When sleep mode is disable it will consume 302 mW - internal battery will last for 23 hours. The above number with an assumption Ethernet, LCD backlight are disable and no usage on 12V onboard output. Could you describe your sensors and measurement interval? Best regards, Rudy Gunawan