Good morning Mike,
Are you connecting to the Digital inputs (1D to 8D)and reading them as a counter or are you connecting to the high speed counters? (1C to 4C)
If you are looking at a contact closure connected to Digital channels 1 to 4 then there is already a debounce circuit so it should not be a problem. The High Speed Counters will count bounces if the input isn't clean.
Cheers,
Roger