Hi,
I'm getting noise of f~ 125kHz and magnitude ~400mV (I've checked it with an oscilloscope) which, as you can imagine is completely drowning out the ~4mV signals from my type K thermocouples when they are at ambient temp.
I have placed the 5V AC power supply in a separate shielded cable - so I know the noise is coming from an external source. So far the only thing that seems to work is making a little RC low pass filter on each channel before I connect it to the dataTaker...
But before I perform this tedious task on all 60-odd channels, I just wanted to check if there happens to be a way to do this through the software (e.g. Delogger).
Thanks.
Rebecca