Voltage measurement on solar module

Please advise before I burn the logger, is it possible to measure the voltage from a solar module using a DT80 directly without putting attenuator?

I'm looking at a voltage of 20-25 V from a module. The manual says one can measure up to 30V. I want to be sure. Please advise.

bm

Please advise before I burn the logger, is it possible to measure the voltage from a solar module using a DT80 directly without putting attenuator? I'm looking at a voltage of 20-25 V from a module. The manual says one can measure up to 30V. I want to be sure. Please advise. bm

Good afternoon Mahambah,

The high Voltage channel type can read up to 30 VDC as per the specification sheet. The HV channel type automatically selects the attenuator. You must have the attenuator selected to read Voltages over 3 VDC.

Cheers,
Roger

Good afternoon Mahambah, The high Voltage channel type can read up to 30 VDC as per the specification sheet. The HV channel type automatically selects the attenuator. You must have the attenuator selected to read Voltages over 3 VDC. Cheers, Roger