Support Forums
100ms trigger schedule gives me onlt 3 readings per second

Hi, i was using this schedule
RA"US"(DATA:OV:1D)100T 1V(A) 1V(A) 2V(A) 2V(A)
i read in the manual that this 100T means it will trigger every 100ms, so i was expecting that i will get around 10 readings in a second. When i extracted the csv file i did not see more than 3 readings in the same second, so i am not sure this is accurate. Also i was expecting the time stamp to be more exact in terms of ms, it only show the time in seconds and then it stops. So what can i do to have the right amount of data shown and extracted?

Hi, i was using this schedule RA"US"(DATA:OV:1D)100T 1V(A) 1*V(A) 2V(A) 2*V(A) i read in the manual that this 100T means it will trigger every 100ms, so i was expecting that i will get around 10 readings in a second. When i extracted the csv file i did not see more than 3 readings in the same second, so i am not sure this is accurate. Also i was expecting the time stamp to be more exact in terms of ms, it only show the time in seconds and then it stops. So what can i do to have the right amount of data shown and extracted?

Hi Gabriel,

If you have series 3 logger, 100 ms rate is not possible for 3 channels because the fastest rate would be 40 ms per 1 channel. But if you have series 4 logger, you can set 100 ms rate (fastest speed 25 ms/ channel).

Please note to get into the fastest speed, you will need to enable few setting, i.e: auto calibration off (\k) or raising the bar for calibration level (P0=5), enable analog circuit (P21=1), increasing frequency filter (P11=1000), reducing channel delay (MD5).

Best regards,
Rudy Gunawan.

Hi Gabriel, If you have series 3 logger, 100 ms rate is not possible for 3 channels because the fastest rate would be 40 ms per 1 channel. But if you have series 4 logger, you can set 100 ms rate (fastest speed 25 ms/ channel). Please note to get into the fastest speed, you will need to enable few setting, i.e: auto calibration off (\k) or raising the bar for calibration level (P0=5), enable analog circuit (P21=1), increasing frequency filter (P11=1000), reducing channel delay (MD5). Best regards, Rudy Gunawan.
40
1
2
live preview
enter atleast 10 characters
WARNING: You mentioned %MENTIONS%, but they cannot see this message and will not be notified
Saving...
Saved
With selected deselect posts show selected posts
All posts under this topic will be deleted ?
Pending draft ... Click to resume editing
Discard draft