Dear Sir,
Pl let me know how to communicate DT85 with Tablet and access dEX Software using (through Chrome or other web browser) Tablet.
What should be the minimum hardware and software configuration for above case?
Regards,
Gpradipta
I have tried to set up a daily schedule to save newly acquired data to a USB memory stick.
I set this up in dEX configuration builder by adding a new schedule to run every day at midnight, then added a manual channel with the command
COPYD dest=A: format=csv archive=N src=BD start=new;
If I issue this command in the command window, it successfully saved data to the USB device.
Wh
renrod
posted Apr 19 '19 at 5:13 am
Dear all,
We use a device from http://www.soilmoisture.com/TDR-Equipment/ (all 3 types) for mesurement 16 channels of water content with time domain reflectometrie system.
How can I control and readout this with a dEX - script at DT85.
Please help me
Hello,
We can't read the legacy configuration back in the data-logger. In order to gain storage space for the installation files we deleted the job, data, alarms, configuration out of the B drive.
Should we create a new job first? Or should we go back to version 9.08 and forget about the new version (where can we get version 9.08 from)? Re-make the configuration which would take us a lot
Some questions about security settings on the DT85:
Are the firmware security settings documented in Part Q of the manual (PROFILE AUTHENTICATION ENABLED, USER, SIGNOFF, SIGNON, WHO) are the same as those set in dEX security?
Can more than one user be logged on at once (e.g. of the DT85 is accessed via an internet connection)? If not, what will a user see if they try and access the DT85 e
Ren Yan
posted Apr 19 '19 at 5:11 am
Hi,
We have always connected our DT85 (V6.22) to PC with DeTransfer or DeLogger using USB successfully. But now it can not be connected, because there has no USB Serial Port display in the Windows Device Manager when insert the USB of logger to the PC.
We tried to change the PC and USB, restart the logger, but still no any response received.
However,when we changed to use a USB to RS
Hello I have a DT85 which will not sleep. It has external power and I have set registers P15=3 and P17=300 but it stays on between 4 hour uploads. I see that if there incomplete jobs entered it will not sleep. How would I see these and how to delete these? Is there another reason for it not going to sleep
I have an Identical machine which is sleeping no problem.
Program as follows any help
Teabone
posted Apr 19 '19 at 5:00 am
I want to run my DT85 from a USB stick. My understanding is that this can be done by porting your program to a file called oninsert.dxc.
I am currently using DeLogger4 to create my job which I can run when connected via USB cable. For the life of me I can't find a way to create an oninsert.dxc file with this program.
Can DeLogger create this file or should I be using a different program?
Hi I have a DT85 which i am logging some temperatures & solar radiance.
I have 3 schedules the first is a standard logging schedule at 15 minutes it simply logs the radiance and temperatures.
the second is the same channel setup except I am logging every second and using CV's to accumulate the total radiance.
The third schedule i wish to setup to log the following.
the average &
Our DT85 inexplicably neglected to log 7 minutes of data on a minute schedule. The last successful data record logged before the data loss had a time stamp of 12:00(noon). I glanced through the service log to see if there was an error around the time of the data loss and I found this:
2014/09/24,12:01:24
An error occurred at buf.c:19
Task: HTTP 2
Current task : HTTP 2
Error loc
Occasionally our DT85 will reset and we'll lose about 7 minutes of data. This shows up on the event log at the time of the data loss:
EVENT ,2014/10/16,12:27:21,0.755615,"No HTTP listener running: restarting"
EVENT ,2014/10/16,12:27:24,0.937377,"Reset 9.10.4890 PLL/WDOG/DEBUG (c860400d)"
I was told that the 9.14 firmware was buggy, therefore I downgraded the firmware to 9.10 thinking t
As part of my student project I'm trying to establish a communication between the DataTaker DT85 and Swegon modbus gateway connected to CASA
ventilation unit (this is the first time I'm working both with modbus protocol and DataTaker). The aim is to have the values measured by the sensors in the unit also logged by the data logger. However, I receive the error E124.
The slave device is pow
Kaltal
posted Apr 19 '19 at 4:13 am
Hi All,
I have a series 3 DT85 v9.14. It as not been used much over the last year, and now just when I need it it seems to be giving me an error. As it collects data it will, all of a sudden, stop, lights flash, and the message SW exception appears.
I was having a problem with the unit running on battery, in that it was losing power then turning back on then losing power. I traced that t
magnum
posted Apr 19 '19 at 4:11 am
Hello -
I am having trouble upgrading the software on a DT85 (I believe series 1). S/N 083710
I normally use DeTransfer for a newer logger I have but the "upgrade firmware" function is grayed out when I connect.
I tried two different USB sticks - one was an 8 Gig which I understand is too big. I also tried a freshly formatted FAT32 256meg stick and it didn't work either. It just reads
Will
posted Apr 19 '19 at 4:12 am
I have some T/RH sensors (Precon HS-2000V) and want to connect them to the DT 85. But I don't know how to connect the four wires.
Keaton
posted Apr 19 '19 at 3:58 am
I need to measure an analog voltage of +/- 5 Vdc. I know I need to set the channel 10:1 attenuator in order to do this. When I configure the input range to '30V Input' on my channel in dEX and save to the logger, I get the error message as seen in the attachment.
I've tried doing the command through the command window in dEX also [9HV(GL30V)]. Then it seems to be accepted, but when I apply a
eric
posted Apr 18 '19 at 4:54 pm
Hi Support,
I am trying to send the follow data to the RS485 port in HEX: 68 04 00 04 08 by sending \104\04\00\04\08
however it appears that only 68 04 04 08 is sent, the 0x00 in the middle was not sent.
Is there a workaround for this?
Thanks very much for the help
BR
Eric
Hi,
I am using the +5V output to power a humidity sensor, I have used the command +5VSW=1 and set the logger up so that the 12V subsystem is continuously powered. However when I run the logger, the +5V works for short scan times (5secs) but for 15min scans the +5V does not switch on and so the sensor gives a bad reading.
Does whether having the USB connected affect whether the +5V SW is
ITS
posted Apr 18 '19 at 4:48 pm
Hi
I am looking to read PT100 sensors and digital inputs from a novus digiral module http://www.novus.com.br/site/default.asp?TroncoID=621808&SecaoID=819191&SubsecaoID=0&Template=../catalogos/layout_produto.asp&ProdutoID=748482
Can you confirm how this can be done with the dt85 sensor port I was looking at something like 1MODBUS(AD2,R1:1,=12CV) to read the digital state a
We want to use datalogger to measure two channels - load cell and strain gage simultaneously, with a schedule such that, the data is recorded on a switch press and not at a fixed time interval.
I was reading the manual and I found that schedules can be triggered on external events, (Pg - 43 of DT-80 Users manual).
I feel a trigger on switch press should be "Trigger on rising transition
I have a sensor TSL235R - LIGHT-TO-FREQUENCY CONVERTER, datasheet - http://www.rhydolabz.com/documents/sensors/TSL235R_LF.pdf
It provides linear response of ambient irradiance to output frequency. How can I measure this output frequency with a datalogger ?
CP
Huge Fluctuations in measurement. I have connected strain Gage and full bridge load cell for measurement.
noticing very high fluctuation in the measurement. If I touch the DAQ body, strain values and all other sensor values whatever connected in DAQ is varying heavily. I have done the ground connection also. still variations in strain measurements is very high.
It is not because of tempe
After downgrading our firmware to 9.10.4890 per Datataker tech support recommendation our DT85 (series 3) is going through resets about every four days causing the loss of data during the reset.
In the event log the reset error is "Reset 9.10.4890 PLL/WDOG/DEBUG (c860400d)".
Currently we have two schedules running; one 15 sec and one 1 min. In the 15 sec. we read 12 analog and 4 modbus c
We have a similar issue with the logger stopping to respond to be read through Modbus TCP/IP.
The logger works fine for 24-48 hours and then stops communicating over the Modbus TCPIP.
The setup we are using a DT85 connected to 16 CEM channels – a total of 400 HV I/O (we are reading 4-20mA pressures transmitters over a 250Ohm resistor).
The DT85 is connected via a freewave radio and
pwarnes
posted Apr 18 '19 at 4:45 am
Dear all,
I have been experimenting using a DT85 Series 2 to measure relative humidity using the Wet/Dry bulb method. I have been taking the measurements using a pair of K type thermocouples - one of which is 'wet.'
I have tried using the routine outline in the FAQ through DeTransfer. The problem is that the calculation in the formula is quite different from that using a psychrometric cha
rheem
posted Apr 18 '19 at 4:49 am
I have contacted technical support in relation to this issue, but this is a more general question regarding current loops.
When you configure a channel for a current loop (as opposed to a current input), without any scaling, the logger should be 'reading' a 0-100% figure. This is related to the 4-20mA range of these current loop sensors. Is that correct?
If I have configured channels for
rheem
posted Apr 18 '19 at 4:48 am
Hello,
I have searched high and low in the manual and FAQ without luck. The DT85 Series 2 has TWO power supply inputs for the switched excitation and shunt resistor. They are labeled exactly the same (EXT * and EXT #).
My question: Are these inputs tied together (is one EXT * electrically equivalent to the other)? Is one power input used for certain analog channels? If I am using switch
Cmin
posted Apr 18 '19 at 4:17 am
Hi all, hope that somebody can help me with this
I'm using analog sensors connected to a Voltage amplifier with a three conductor cable, one to signal out, one to ground and the third for power supply. The amplifier has an terminal to connect the power supply input +DC voltage (Vcc 5-24 VDC), but is not clear for me to which terminal in the datataker would correspond the power supply connecti
I hope everyone is doing well.
I have the DT85 series 3 dataTaker and I would like to hook up string pots to it for our testing lab. I have no experience with the data taker and would love any help anyone can provide
I've emailed tech support and they suggested that it might by a voltage divider pot but I'm sure that it reads resistance. I even placed a multi-meter across the green and
cbasler
posted Apr 18 '19 at 12:52 am
Hello,
I want to run the DT85 via USB, so therefore I need a USB-Driver. Unfortunately, the Driver on the page seems to be missing. Also, the dEX software platform, which I thought I could find is leading me to an "not found" page. Could you help me with that problems?
Greetings,
Clemens Basler